by WesRagle » Mon Jul 05, 2021 7:11 pm
Hi Maurice,
I'll give this a shot.
It's likely that the "problem" you are seeing is the results of poor reference junction compensation (cold junction compensation) by the RDAC. If you ask a scientist how a TC works he/she will tell you that the emf generated by a TC is a function of the line integral of the temperature gradient along the length of thermocouple wire. If you ask engineers they will likely tell you the voltage generated is a function of T1 - T2 where T1 is the temperature at the sensing (hot) point and T2 is the temperature where the TC voltage is measured.
All of the equations and tables which are used to correlate the measured voltage with the temperature at the tip of the TC assume a T2 of zero degrees centigrade, which of course is rarely the case. So ..., enter reference junction or CJ compensation. The data acquisition system must measure the temperature where the TC joins copper and use that temperature to adjust the measured voltage before using that voltage to calculate the temperature reading. It sounds like the RDAC is doing a poor job of that.
The point is that, assuming a properly implemented system, you can take a blow torch to a point between the sensing point and the measurement point and it will not effect the T/C reading until the T/C is shorted, burned in two, or the extreme temperature soaks to the TC wire end thus effecting T1 .
However, I don't understand how they were able to adjust/miss the reading by 100 Deg.
Edit:
BTW, you didn't use copper extension wire on your TCs did you?
Wes
Wes Ragle
Onex #89
Conventional Gear
Long Tips
Hummel 2400 w/Zenith Carb
Prince P Tip 54x50
First Flight 06/23/2020
42.8 Hrs. as of 10/30/21