Thinking of a TC as everything from the junction of dissimilar metals to the point at which those metals join copper, anything that changes the temperature at either end of the TC will change the measured voltage.
The reason I didn't include T2 in the statement,
The point is that, assuming a properly implemented system, you can take a blow torch to a point between the sensing point and the measurement point and it will not effect the T/C reading until the T/C is shorted, burned in two, or the extreme temperature soaks to the TC wire end thus effecting T1 .
is because in a carefully implemented system T2 is always a known quantity and the measured voltage is always corrected to the known T2.
Understand that careful implementation of CJ compensation is expensive and generally painful. In my professional life we used circular connectors (cannon plugs) almost exclusively. While measuring TCs not only was the cabling TC wire, the pins in the connector were specially machined TC pins. On the other side of the connector was more high quality TC wire which terminated at the PC board right next to an accurate temperature sensor. All voids around the board connection were filled with "copper pour" and the whole mess was potted. All that in attempt to know T2 to 0.1 Centigrade degree. Call it pride, call it specmanship, call it whatever, we went to extremes so the brochures could say we were accurate to within 1 Deg. C.
Smaller, less expensive systems can't/don't/won't go to all that trouble.
I guess the "real question" is : "Where is the CJ temperature measured?". Do they have a sensor inside the box? Is it out on the connector? Do they use the OAT sensor? How do they know the temperature right at the point where the TC wires transition to copper?